calibration of voltage

vgy posted on 26 Oct 2012 10:58 PM PST
Top Contributor
133 Forum Posts

 hi to all,

              In  the design of blood glucose meter , the current from test strip is modified to voltages using TIA , but after that how to calibrate that voltages interms of Glucose concentration?




Re: calibration of voltage

H L posted on 26 Oct 2012 12:59 AM PST
Top Contributor
679 Forum Posts

How about this 

http://www.eetimes.com/design/test-and-measurement/4204495/Blood-Glucose-Meter-Design



Re: calibration of voltage

danaaknight posted on 26 Oct 2012 03:23 AM PST
Top Contributor
1773 Forum Posts

Not being an expert in this field, but don't the test strip manufacturers

have a datasheet for the sensor characteristics ? Which would lead

you to concentration calculations ?

 

Regards, Dana.



Re: calibration of voltage

danaaknight posted on 26 Oct 2012 03:31 AM PST
Top Contributor
1773 Forum Posts

This might help -

 

http://electrochem.cwru.edu/encycl/art-g01-glucose.htm

 

Regards, Dana.



Re: calibration of voltage

H L posted on 26 Oct 2012 05:36 AM PST
Top Contributor
679 Forum Posts

 As dana mentioned, you should get the data form the company making the strip as each company's sensing method would be different.



Re: calibration of voltage

Bob Marlowe posted on 26 Oct 2012 05:49 AM PST
Top Contributor
1768 Forum Posts

There are normalized glocose test-liquids to calibrate those equipments.

 

Bob



Re: calibration of voltage

danaaknight posted on 26 Oct 2012 06:09 AM PST
Top Contributor
1773 Forum Posts

There are test strips for that purpose, with known concentrations

on them.

 

http://www.americandiabeteswholesale.com/catalog/diabetic-test-strips_54.htm

 

Regards, Dana.



Re: calibration of voltage

Bob Marlowe posted on 26 Oct 2012 08:34 AM PST
Top Contributor
1768 Forum Posts

Another challenge:

Often the strips are bar-code-marked to have them calibrated (or date-checked). The code is red off and acted on accordingly.

Within my glucose-meter (Accucheck) is a bar-code reader integrated that reads off the barcode from the strip-container which is a small drum containing 17 strips that is inserted into the meter.

 

Bob



Re: calibration of voltage

PSoC Rocks posted on 28 Oct 2012 05:40 AM PST
Top Contributor
128 Forum Posts

If you are looking for calibrating the internal TIA/signal chain, then take a look at this Application Note - AN68403 (http://www.cypress.com/?rID=50320)






ALL CONTENT AND MATERIALS ON THIS SITE ARE PROVIDED "AS IS". CYPRESS SEMICONDUCTOR AND ITS RESPECTIVE SUPPLIERS MAKE NO REPRESENTATIONS ABOUT THE SUITABILITY OF THESE MATERIALS FOR ANY PURPOSE AND DISCLAIM ALL WARRANTIES AND CONDITIONS WITH REGARD TO THESE MATERIALS, INCLUDING BUT NOT LIMITED TO, ALL IMPLIED WARRANTIES AND CONDITIONS OF MERCHANTABILITY, FITNESS FOR A PARTICULAR PURPOSE, TITLE AND NON-INFRINGEMENT OF ANY THIRD PARTY INTELLECTUAL PROPERTY RIGHT. NO LICENSE, EITHER EXPRESS OR IMPLIED, BY ESTOPPEL OR OTHERWISE, IS GRANTED BY CYPRESS SEMICONDUCTOR. USE OF THE INFORMATION ON THIS SITE MAY REQUIRE A LICENSE FROM A THIRD PARTY, OR A LICENSE FROM CYPRESS SEMICONDUCTOR.

Content on this site may contain or be subject to specific guidelines or limitations on use. All postings and use of the content on this site are subject to the Terms and Conditions of the site; third parties using this content agree to abide by any limitations or guidelines and to comply with the Terms and Conditions of this site. Cypress Semiconductor and its suppliers reserve the right to make corrections, deletions, modifications, enhancements, improvements and other changes to the content and materials, its products, programs and services at any time or to move or discontinue any content, products, programs, or services without notice.