implementation of bio medical signals recording using PSoC
Moderator:
ARVI

Post Reply
Follow this topic



implementation of bio medical signals recording using PSoC

nalini posted on 18 Jun 2012 4:22 AM PST
Senior Member
11 Forum Posts

hi sir, i have doubt regarding the leds using to measure the oxygen concentration in human bieng , we design the sytem with a clip ,on  one side of the clip we place a two leds. My question is  why only we use red, infrared leds. why not other leds. What is the technical reason behind that .




Re: implementation of bio medical signals recording using PSoC

danaaknight posted on 18 Jun 2012 05:40 AM PST
Top Contributor
1773 Forum Posts

You need to use a RED and a INFRARED led in design.

This should help - http://www.oximetry.org/pulseox/principles.htm

 

Regards, Dana.



Re: implementation of bio medical signals recording using PSoC

nalini posted on 19 Jun 2012 10:49 PM PST
Senior Member
11 Forum Posts

thanx sir , which information u gave to me is very useful. thank u



Re: implementation of bio medical signals recording using PSoC

nalini posted on 19 Jun 2012 10:49 PM PST
Senior Member
11 Forum Posts

thank u sir



Re: implementation of bio medical signals recording using PSoC

nalini posted on 21 Jun 2012 03:56 AM PST
Senior Member
11 Forum Posts

hi sir,

where we can find the one wire protocal datasheet in psoc designer 4.4 version. how can we interface ds18b20 with onewire protocal.



Re: implementation of bio medical signals recording using PSoC

nalini posted on 18 Jul 2012 08:02 PM PST
Senior Member
11 Forum Posts

hi sir,

           what is the main differance betweeen PSoC and SoC . And what are the advantages of  PSoc compared to soc.



Re: implementation of bio medical signals recording using PSoC

arvi posted on 18 Jul 2012 09:53 PM PST
Cypress Employee
119 Forum Posts

 Hi nalini,

 

You can find information about the 1-Wire interface and how you can use DS18S20 temp sensors here > AN2163 - PSoC® 1 - Interfacing to 1-Wire/Two-Wire Digital Temperature Sensors using PSoC® 1

 

Regards,

Arvind



Re: implementation of bio medical signals recording using PSoC

Bob Marlowe posted on 19 Jul 2012 10:20 PM PST
Top Contributor
1768 Forum Posts

The difference between PSoCs and the rest of the embedded world are analog components as comparqators, oppamps, TIAsm sample & holds, track & holds DACs and so on.

In "normal" systems is very often a requirement for signal-conditioning as amplification, filtering etc. With a PSoC you can get a one-chip solution with only a few additional discrete components (if any) making developement, layout. part-count and pcb easier and smaller. A real compülete programmable system on a sole chip.

 

Bob  



Re: implementation of bio medical signals recording using PSoC

Bob Marlowe posted on 19 Jul 2012 10:56 PM PST
Top Contributor
1768 Forum Posts

I just see (overlooked it)

Designer 4.4 is a rather old version, there have been several updates and corrections yet. I would suggest you download (for free of course) Designer 5.2 from the Cypress website and get familar with it. For older projects it might be necessary to download and install the 5.0 version as well (they are co-existent) to update your projects from versions 4 to 5.

 

Bob



Re: implementation of bio medical signals recording using PSoC

H L posted on 19 Jul 2012 11:29 PM PST
Top Contributor
679 Forum Posts

A lot of other chips do have programable peripherials on chip. But PSOC more like having a reasonable size CPLD and the possibility of routing every type of signal to every pin ( nearly 100 %) makes it better then other chips.



Re: implementation of bio medical signals recording using PSoC

nalini posted on 19 Jul 2012 11:53 PM PST
Senior Member
11 Forum Posts

hi sir,

                       Thank you for your information and sir can you clear my doubt amplifiers and filters. in psoc we have programable filters and amplifiers. hOw the gain of amplifier is programed and how the frequency responce of filter is programed.



Re: implementation of bio medical signals recording using PSoC

H L posted on 19 Jul 2012 12:19 AM PST
Top Contributor
679 Forum Posts

For filters:

You pick and place the filter into your designer and right click on the module, you will find a filter design wizard where you can change the type, gain, frequency and sampling clock for your design.
For amplifiers.:
You have to place the amplifier into the designer first and then change the properties.
I think you should download some videos from Cypress to familiar with the operation of the designer.

:-)



Re: implementation of bio medical signals recording using PSoC

Bob Marlowe posted on 19 Jul 2012 01:24 AM PST
Top Contributor
1768 Forum Posts

...or download and install PSoC Creator 2 and give it a try. Follow the advice to watch the videos, using Creator is really easy!

 

Bob



Re: implementation of bio medical signals recording using PSoC

Bob Marlowe posted on 19 Jul 2012 01:31 AM PST
Top Contributor
1768 Forum Posts

oops, sorry, my fault: not the Creator, that's for PSoC3 and 5 (although they are great as well) for PSoC1 you'll need Designer 5.2

 

Bob



Re: implementation of bio medical signals recording using PSoC

danaaknight posted on 19 Jul 2012 03:19 AM PST
Top Contributor
1773 Forum Posts

For filter, as previously mentioned, use the wizard, or the Excel

design sheets located (Designer 5.2) in menu item "Help" >

"Documentation" > "Filter Design".

 

For videos and training go to http://www.cypress.com/?id=1162&source=header

 

Regards, Dana.



Re: implementation of bio medical signals recording using PSoC

arvi posted on 19 Jul 2012 03:34 AM PST
Cypress Employee
119 Forum Posts

  Hi nalini,

 

Can you tell me more about your application? Usually Instrumentation amplifiers are recommended for medical designs.

They are known as INSAMP in PSoC Designer and can be found under the "Amplifiers" category in User module window.

You will be able to read differential signals as well as amplify them (Gain = 2x to 16x) using INSAMP. Alternatively you can also use a programmable gain amplifier (PGA) with gain upto 48 times input signal.

You can find an example implementation of the INSAMP here > http://www.cypress.com/?id=4&rID=58579

 

Regards,

Arvind



Re: implementation of bio medical signals recording using PSoC

nalini posted on 20 Jul 2012 07:21 AM PST
Senior Member
11 Forum Posts

hi sir ,

            i am working on medical application, as u said INSAMP is used and along with that i am using  the modules in  the attached document what i had pin here. here i am measuring some parameters to check health condition of a patient like temperature, heart beat, oxygen concentration, ph level in blood, ecg.



Re: implementation of bio medical signals recording using PSoC

danaaknight posted on 20 Jul 2012 10:52 AM PST
Top Contributor
1773 Forum Posts

A Instrumentation amplifier is used when you have high common mode signals you need to

reject. Like EKG rejecting 50/60 cycle common mode pickup.

 

Attached is a good reference.

 

Regards, Dana.

 

 



Re: implementation of bio medical signals recording using PSoC

danaaknight posted on 20 Jul 2012 10:53 AM PST
Top Contributor
1773 Forum Posts

Another good reference.

 

Regards, Dana.



Re: implementation of bio medical signals recording using PSoC

danaaknight posted on 20 Jul 2012 10:54 AM PST
Top Contributor
1773 Forum Posts

And one more reference.

 

Regards, Dana,



Re: implementation of bio medical signals recording using PSoC

nalini posted on 29 Jul 2012 12:56 AM PST
Senior Member
11 Forum Posts

hi sir,

          i am facing a problem with the PSoC 1,  4.4 version software. i installed it in proper way. but while i am doing simple program like lcd display, when i go through application editor  phase after interconnection phase i am getting the window as shown in attached document.the main c program is not coming. sir could you tell me what is that problem.



Re: implementation of bio medical signals recording using PSoC

Bob Marlowe posted on 29 Jul 2012 01:15 AM PST
Top Contributor
1768 Forum Posts

Only very few people here have access to the (rather old) version 4.4. Why not update to 5.2 and give it a try? As far as I know both versions are co-existent, although you will not be able to open a 4.4 project with 5.2, I think you have to use an intermediate version of 5.0 to accomplish that.

 

Bob



Re: implementation of bio medical signals recording using PSoC

arvi posted on 29 Jul 2012 01:23 AM PST
Cypress Employee
119 Forum Posts

 

Nalini,

 

The LCD usermodule will not show up on the interconnect view as it will not consume any digital blocks. Which means that you cannot place it. 

 

After selecting it using Config > Selection, when you do a Config > Generate Application you will be able to see the LCD libraries generated.

 

To view these files, select the Application Editor button and expand the library source files list: (see attached image)

 

 

For more information please go through the IDE User Guide in C:\Program Files\Cypress MicroSystems\PSoC Designer\Documentation\

 

I'm repeating this again, but my sincere recommendation is that you upgrade to the latest version of PSoC Designer (v5.2) software. You will a huge improvement in terms of ease of use and more number of user modules and devices supported.

 

-Arvind



Re: implementation of bio medical signals recording using PSoC

danaaknight posted on 29 Jul 2012 03:54 AM PST
Top Contributor
1773 Forum Posts

The LCD module does have to be "place"-ed, to configure it mode and choose its

port. It does not, however, show up in routing view.

 

Regards, Dana.



Re: implementation of bio medical signals recording using PSoC

nalini posted on 29 Jul 2012 05:54 AM PST
Senior Member
11 Forum Posts


k sir here i am telling the procedure what i am following. first i click psoc designer, and then start new project and there i am giving the name of the project , then cliking on next and then generate main file using c, after that finish. next i am following user module selection  in that i select lcd , next interconnection in that i am giving port 2 and then application editor the following documents shows the procedure.



Re: implementation of bio medical signals recording using PSoC

Bob Marlowe posted on 29 Jul 2012 06:13 AM PST
Top Contributor
1768 Forum Posts

Doesn't look too bad, next thing would be to "build" your project which will first generate the neccesary bits to set your hardware porperties and connections and second compile all the c sources (some of them have been generated during the first step).

Again: Can you tell us ANY reason why you are relucant to download and install Designer 5.2 which not only has corrected a bunch of errors but also comes with new versions of all the usermodules.

 

Bob



Re: implementation of bio medical signals recording using PSoC

Bob Marlowe posted on 29 Jul 2012 06:16 AM PST
Top Contributor
1768 Forum Posts

Necessary, necessary, necessary, I think I'll never get that right!!!






ALL CONTENT AND MATERIALS ON THIS SITE ARE PROVIDED "AS IS". CYPRESS SEMICONDUCTOR AND ITS RESPECTIVE SUPPLIERS MAKE NO REPRESENTATIONS ABOUT THE SUITABILITY OF THESE MATERIALS FOR ANY PURPOSE AND DISCLAIM ALL WARRANTIES AND CONDITIONS WITH REGARD TO THESE MATERIALS, INCLUDING BUT NOT LIMITED TO, ALL IMPLIED WARRANTIES AND CONDITIONS OF MERCHANTABILITY, FITNESS FOR A PARTICULAR PURPOSE, TITLE AND NON-INFRINGEMENT OF ANY THIRD PARTY INTELLECTUAL PROPERTY RIGHT. NO LICENSE, EITHER EXPRESS OR IMPLIED, BY ESTOPPEL OR OTHERWISE, IS GRANTED BY CYPRESS SEMICONDUCTOR. USE OF THE INFORMATION ON THIS SITE MAY REQUIRE A LICENSE FROM A THIRD PARTY, OR A LICENSE FROM CYPRESS SEMICONDUCTOR.

Content on this site may contain or be subject to specific guidelines or limitations on use. All postings and use of the content on this site are subject to the Terms and Conditions of the site; third parties using this content agree to abide by any limitations or guidelines and to comply with the Terms and Conditions of this site. Cypress Semiconductor and its suppliers reserve the right to make corrections, deletions, modifications, enhancements, improvements and other changes to the content and materials, its products, programs and services at any time or to move or discontinue any content, products, programs, or services without notice.

Spec No: None; Sunset Owner: GRAA; Secondary Owner: RAIK; Sunset Date: 01/01/20